Skip to main content
Examonaut

Topic 9A · Lesson 3 of 4

Instrumental methods and flame photometry

Instrumental methods

Chemical tests are cheap and simple, but they need a reasonable amount of sample and depend on human judgement of colours. Instrumental methods use machines, and have three big advantages:

  • sensitive: they can detect very small amounts of a substance;
  • accurate: they give precise, reliable measurements rather than an opinion about a colour;
  • rapid: many samples can be analysed quickly and often automatically.

They can also measure how much is present, not just what is present. The disadvantages are that the equipment is expensive, and needs trained people and regular calibration.

Flame photometry

A flame photometer sprays a solution into a flame and analyses the light given out.

  • Identifying ions: each metal ion gives out light at its own set of wavelengths (its emission spectrum). Comparing the lines in a sample's spectrum with reference spectra identifies the metal ions present, even in a mixture, as shown below.
  • Measuring concentration: the intensity of the light increases with the concentration of the ion. You measure the intensity for several standard solutions of known concentration, plot a calibration curve, then measure the unknown and read its concentration from the graph.
lithiumsodiumpotassiumsample400500600700800wavelength / nm (simplified)

The sample's spectrum has lines at 404, 589 and 766 nm. The 589 nm line matches sodium and the 404 and 766 nm lines match potassium, so the sample contains sodium and potassium ions but no lithium. A flame test by eye would probably have shown only yellow, because sodium's colour hides potassium's lilac.

Worked example

Using a calibration curvetop grade

Standard sodium solutions of 0, 2, 4, 6 and 8 mg dm⁻³ give flame photometer readings of 0, 20, 40, 60 and 80 (illustrative readings), as shown below. A sample of river water is diluted 10 times, and the diluted sample gives a reading of 46. Find the sodium ion concentration in the river water.

Show solution
02468020406080sodium concentration / mg dm⁻³emission reading
  1. The calibration line is straight: a reading of 10 for every 1 mg dm⁻³.
  2. A reading of 46 corresponds to 46÷10=4.646 \div 10 = 4.6 mg dm⁻³ in the diluted sample.
  3. The river water was diluted 10 times, so the original concentration is 4.6×10=464.6 \times 10 = 46 mg dm⁻³.
  4. Why dilute? The undiluted reading (460) would be far off the end of the calibration line, and you should only read concentrations within the range of your standards.

Practise this topic