Measuring and drawing cells

Core practical 2: using a light microscope to measure and draw cells

Aim. To measure the size of cells in a specialised tissue using a calibrated eyepiece graticule, and to make a labelled scientific drawing.

Apparatus. A light microscope with ×10 eyepiece and ×4, ×10 and ×40 objectives; an eyepiece graticule (a transparent disc with a scale of, for example, 100 equal divisions, placed in the eyepiece); a stage micrometer (a slide with an accurately ruled scale, often 1 mm divided into 100 divisions of 10 µm); prepared or freshly made slides (for example, onion epidermis stained with iodine, or a cheek cell smear stained with methylene blue); and a pencil and plain paper.

Why calibrate? The graticule's divisions are arbitrary. Their real size depends on the objective lens being used: at higher magnification each division covers a smaller distance on the specimen. So you must work out what one eyepiece unit (epu) is worth for each objective, using a scale of known size.

Method.

  1. Step 1: Set up the microscope with the lowest-power objective. Focus on the stage micrometer using the coarse focus, then the fine focus.
  2. Step 2: Switch to the objective you will use for measuring. Rotate the eyepiece so that the two scales lie side by side, and line up the zero marks.
  3. Step 3: Find a point further along where a division of the graticule lines up exactly with a division of the stage micrometer, as in the diagram below. Record both readings.
  4. Step 4: Calculate the length of one eyepiece unit.
  5. Step 5: Replace the micrometer with your specimen, without changing the objective. Measure several cells in eyepiece units, along the same axis, and convert to µm. Find a mean.
  6. Step 6: Recalibrate if you change to a different objective lens.
eyepiece graticule (arbitrary units)0102030400510152025stage micrometer (each division 10 µm)the scales line up at both ends

Worked example

Calibrating and measuring

In the diagram above, using the ×40 objective, 40 eyepiece units line up exactly with 25 stage micrometer divisions. Each stage division is 10 µm. A cell measures 12 eyepiece units long. Calculate its length.

Show solution
  1. Length of the stage scale: 25×10=25025 \times 10 = 250 µm.
  2. Value of one eyepiece unit: 250÷40=6.25250 \div 40 = 6.25 µm.
  3. Length of the cell: 12×6.25=7512 \times 6.25 = 75 µm.

Worked example

Mean size with uncertaintytop grade

Five cells measured 11, 12, 12, 13 and 14 eyepiece units, with 1 epu = 6.25 µm. Calculate the mean length and comment on the precision of the method.

Show solution
  1. Mean in eyepiece units: (11+12+12+13+14)÷5=62÷5=12.4(11 + 12 + 12 + 13 + 14) \div 5 = 62 \div 5 = 12.4 epu.
  2. Convert: 12.4×6.25=77.512.4 \times 6.25 = 77.5 µm.
  3. Precision: each reading can only be made to the nearest whole division, so the uncertainty in each is about ±0.5 epu, or ±3 µm. As a percentage of a 12 epu reading this is 0.5÷12×100≈4%0.5 \div 12 \times 100 \approx 4\%. Measuring more cells, and using a higher-power objective so each cell covers more divisions, would reduce the effect of this uncertainty.

Making a scientific drawing. The example below shows the conventions examiners look for.

  • Use a sharp HB pencil and plain paper; draw large (at least half the space given).
  • Use single, clear, continuous lines. No shading or colouring.
  • Draw only what you see, keeping the proportions correct. Draw a small number of cells (often three to six) in detail rather than many roughly.
  • Label with ruled lines that touch the structure, do not cross each other and have no arrowheads. Labels go to the side, written horizontally.
  • Add a title and the magnification (or a scale bar).
nucleuscell wallcytoplasmOnion epidermis cells, ×400sharp single lines, no shading, ruled labelsthat do not cross; drawn large

Risks. The ethanol in some stains is flammable; iodine and methylene blue stain skin and clothes. Never use direct sunlight as a light source with a mirror, because it can damage your eyes. Handle glass slides and cover slips carefully.

Errors and improvements. Cells may be cut or squashed at an angle, so measure many cells and use the mean. Lining up scales by eye introduces small errors, so choose a point where the lines coincide exactly, as far along the scale as possible. A digital camera with image-analysis software can measure more precisely.