Counting cells with a haemocytometer

Measuring growth: the four methods

Biologists need to know how many cells a culture contains, for example to follow a growth curve or to compare antibiotics. Bacteria are tiny and numerous (often over 10810^8 cells in each cm³ of broth), so no method counts every cell directly. There are four approaches, summarised in the diagram below. The most important distinction is between a total count, which includes dead cells, and a viable count, which counts only living cells that can still divide.

Haemocytometer countcells counted in a grid ofknown volume under microscopeTOTAL countDilution platingdilute, spread on agar,count coloniesVIABLE countDry massfilter, dry to constantmass, weighTOTAL (mass)Turbiditycolorimeter measuresabsorbance of brothTOTAL countOnly dilution plating counts just living cells

Direct cell counts with a haemocytometer

A haemocytometer is a thick microscope slide with a grid etched into it. When a coverslip is placed on top, the space between the grid and the coverslip has a fixed depth, usually 0.1 mm. So every square of the grid holds a known volume of liquid. You count the cells in some squares under the microscope and scale up.

The diagram below shows part of the grid. In the most common type, the middle 1 mm × 1 mm square is divided into 25 medium squares, each 0.2 mm × 0.2 mm.

one mediumsquare:0.2 mm × 0.2 mmdepth undercoverslip:0.1 mmcount cellstouching top andleft edges onlypart of the haemocytometer grid

How to do it:

  1. Step 1: Find the volume of one square: length × width × depth. Keep everything in mm, then convert: 1 mm3=10−3 cm31 \text{ mm}^3 = 10^{-3} \text{ cm}^3.
  2. Step 2: Count the cells in several squares chosen at random and find the mean per square. For cells that touch the edges, count those on two sides (for example top and left) and ignore those on the other two, so no cell is counted twice.
  3. Step 3: Cells per cm³ = mean number per square ÷ volume of one square in cm³.
  4. Step 4: Multiply by any dilution factor used to make the sample thin enough to count.

A haemocytometer gives a total count, because dead cells look the same as living ones. Adding a stain such as trypan blue fixes this: it enters only dead cells, whose membranes are damaged, so you can count the unstained (living) cells separately.

Worked example

A haemocytometer count

A yeast culture was diluted 1 in 10. On a haemocytometer with a depth of 0.1 mm, five medium squares (0.2 mm × 0.2 mm) contained 12, 9, 11, 10 and 8 cells. Calculate the number of cells per cm³ in the original culture.

Show solution
  1. Volume of one medium square: 0.2×0.2×0.1=0.004 mm30.2 \times 0.2 \times 0.1 = 0.004 \text{ mm}^3. Convert to cm³ because the answer is per cm³: 0.004×10−3=4×10−6 cm30.004 \times 10^{-3} = 4 \times 10^{-6} \text{ cm}^3.
  2. Mean count: (12+9+11+10+8)÷5=10(12 + 9 + 11 + 10 + 8) \div 5 = 10 cells per square. Using a mean of several squares reduces the effect of cells being unevenly spread.
  3. Cells per cm³ of the diluted sample: 10÷(4×10−6)=2.5×10610 \div (4 \times 10^{-6}) = 2.5 \times 10^{6}.
  4. The sample was diluted 1 in 10, so the original had ten times more: 2.5×1072.5 \times 10^{7} cells cm⁻³.